troubleshooting and repair of a double-beam-photometer
OP #7428689
Hello,
it is about troubleshooting and repair of a double-beam-photometer.
The photometer has three failures.
  1. the measurement value is quite strongly fluctuating.
  2. the measurement value (without sample) should be E = 0.000 but instead of it quite strongly fluctuates around E = 0.300 (E = 0.24_ - 0.35_).
  3. when measuring samples: the higher the concentration the lower is the measured value, the opposite of what it normally should be!
First of all I am only interested in solving failure number one!
To locate the failure I need advice because I am not an electronics-expert!
  1. circuit diagram: https://www.file-upload.net/download-15154708/measurement-value-generation.jpg.html
Question 1: Is the direct (output voltage) signal from the photomultiplier tube a correct pmt-signal? 2. pmt-output-voltage-signal: https://www.file-upload.net/download-15154693/pmt-output-signal.mp4.html setting: 50 mV/div and 5 ms/div.
The photometer normally should have a rectangular signal of U = +0.6 V for E = 0 but it strongly fluctuates around 2 V - 2.5 V (for E ≈ 0.300). 3. signal: https://www.file-upload.net/download-15154698/signal.mp4.html
  1. guide-voltage: https://www.file-upload.net/download-15154705/-4V_guide_voltage.mp4.html
  2. pmt-high-voltage: https://www.file-upload.net/download-15154706/-1000V_high_voltage.mp4.html
I am no electronics-expert but I think possible reasons for fluctuation of the signal/measuring-value could be:
  1. unstable guide-voltage
  2. unstable pmt-high-voltage
  3. pmt-defect
  4. other
Thank you!
(Company: matzetronics) #7428714
First you should find an oscilloscope and a high voltage voltmeter. As a first step make sure that all power supply lines give constant DC voltage without fluctuations. Measure the PMT supply with the high voltage voltmeter or a HV probe for the oscilloscope. Continue by verifying the waveforms shown in the diagram and measure the voltages shown. Start with the frontend amplifier and continue with the Sample & Hold stages and the lin/log converter.
Pin 6 of the Clock Circuit (CC) should carry the signal from the interupter wheel and it should be close to a square wave. Without this signal the whole circuit will not work properly. It clocks the S&H switches and so the amplifiers MSA1 and MSA2.
OP #7429467
Hello Matthias,
thank you very much, your advice has already helped me much!
I have got an analogue-oscilloscope with a HV/HF-probe up to U = 2500 V but no high-voltage-voltmeter. There is a table of almost all possible values that I measured yet and I am going to remeasure some values first of all particularly pin 6 of CC because the first value of pin 6 CC that I looked up is U = 0.4 - 0.6 V!
Last time the video-links did not work therefore here are new working links!
Videos of clock-circuit (CC)-signal measured at pin 6: settings: • 0.2 V/div and 5 ms/div • 0.2 V/div and 2 ms/div • 0.2 V/div and 1 ms/div • 0.5 V/div and 5 ms/div https://workupload.com/archive/7G9LeYrjjH
Question 2: Is this a good signal? I don't think so!
  1. pmt-output-voltage-signal: setting: 50 mV/div and 5 ms/div.
The photometer normally should have a rectangular signal of U = +0.6 V for E = 0 but it strongly fluctuates around 2 V - 2.5 V (for E ≈ 0.300). 3. signal: setting: 1 V/div and 5 ms/div
  1. guide-voltage: setting: 1 V/div and 5 ms/div
  2. pmt-high-voltage: setting: 200 V/div and 5 ms/div https://workupload.com/archive/KPTzue9YDX
Thank you!
(Company: matzetronics) #7430119
Looking at the circuit this is anything but a good output from the phototransistor. The level is much too small and there's a signal with it which doesn't look like the expected clean squarewave.
To see if its a issue with the supply of the CC, it might be a good idea to stop the motor and turn the interrupter by hand if that is possible. The output should jump between high and low. No light except the LED should light th phototransistor. Use the oscilloscope to verify the supply voltages of the CC and make sure it's a clean DC voltage, for both +15 and -15V and of course the 5V.
If the output of the phototransistor is still this low check the LED. Older LEDs tend to loose brightness over time and it might be the case here. If possible, use another lightsource next to the LED while measuring Pin 6.
If the unit is a complete forked photoelectric sensor, it might be best to replace it.
OP #7432899
Hello Matthias,
first of all thank you very much once again!
I am still busy with remeasuring and to put in order the data material and writing it down.
I will report to you as soon as possible. Probably tomorrow.
OP #7435537
Hello Matthias,
I am back again; sent you a personal message.
Here are a few remeasured values.
The operational voltages of the clock-circuit (CC) with different settings: pin 16: U = -15 V
pin 18: U = +15 V
pin 2: U = +5 V
pin 25: U = -15 V
pin 26: U = +5 V
pin 31: U = +5 V
pin 6: presumably phototransistor
https://workupload.com/archive/WwJ5hdHdhL
Unfortunately I have forgotten to make films of pin 26! They will follow soon.
(Company: matzetronics) #7435547
Its not very helpful to show movies of an oscilloscope only showing a straight line. If the voltage has no humming or other noises its ok. I suspect that you are not very familiar with electronics but you have to find something near you who is. The usage of photometer leads me to think that you are in a scientific environment and surely there must be someone who can help on location.
What you really need to find out is if the level of the phototransistor output is sufficient to clock the Clock Circuit. And i can't identify any phtotransistor in the supplied picture. There must be a motordriven interrupter wheel and an LED/Phototransistor combination which is interrupted by the wheel. This signal is used to clock the amplifiers and sample&hold stages with the help of the Clock Circuit CC.
OP #7435951
Hello Matthias,
thank you very much. Until now you have helped me very much yet and I know what is to do next.
Sorry, I didn't want to bore you but I didn't know if the slight fluctuation of the line is normal. From now on I know it.
I am an electronics newcomer. That is not so easy.
That is right. But even that is very difficult.
Ok I have fully understood. I have to find the LED/phototransistor combination and must then measure the phototransistor output level. This is the absolutely essential next step for troubleshooting!
The interrupter wheel/chopper (light beam divider) is in the optical system case. Here are two photos of it: https://workupload.com/archive/9f9RssHQR9
The phototransistor should be near the chopper case. I hope I will find the phototransistor because it is difficult to look and handle in the optical system case!
#7435981
Ph1 is direct under La1 and Motor M. Signal is on Line 6, it has probably high impedance, so use a 10:1 probe when measuring with oszilloscope, even if signal level is small. Ground level is on line 1 and 23. Signal is also available on KL4-1, Gnd on KL4-2.
If the clock-circuit is working, there must be the chopper signal on Line 8, 10, 11 and 12 with digital level.
OP #7436525
Hello Wolf,
thank you very much for your help!
Matthias means photo by picture that I have sent here: pin 6: presumably phototransistor
Now I have finally understood that by La 1 (lamp 1) is meant a LED and namely the one LED which feeds phototransistor Ph 1 with light pulses that are generated by the chopper.
The next thing I have to do today is to search for KL 4, La 1 and Ph 1; hope I will find them!
When I first studied the circuit diagrams I didn't know what meant "Ph 1". I couldn't find the circuit symbol because it is an outdated one. A little bit later it was clear to me that it must be a phototransistor but its function wasn't clear. Thanks to Matthias I know now!
No idea I had that my photometer needed a phototransistor to work. I have always thougt:
light source ---> chopper chops/divides the light beam ---> photomultiplier tube ---> signal amplification and processing ---> output to display ---> ready! (process is highly simplified and abbreviated)
I have to say that I understand my machine better and better.
(Company: matzetronics) #7436550
Yeah, i saw that but can't find any phototransistor there. I see a measuring point at the end of a 2k21 resistor. The funny loop is for clipping a probe onto it.
Those must be close to the chopper motor where the light beam of the LED is interrupted by some kind of iris or so. The base clock of this signal is then used to switch between the two MSA blocks and is also responsible for switching the corresponding Sample & Hold circuits. in the meantime between measurements this circuit holds the last sampled value for evaluating. As this is a double beam photometer, the circuits are used in an alternating mode, one circuit for each beam, selected by the chopper.
#7436587
To find KL4, search for the motor, probably you can hear him working. Follow the shaft until you find the rotating segmented shutter wheel. Look for the 68R resistor, it is lokated from KL4-4 to KL4-5. The voltage at the 68R should be about 3Vdc.
According to the schematics, they have used a LED as lightsource. Look if you can see some (red?) light if it is dark in the room. Maybe they use a invisible IR LED. Maybe they use a beam enclosure or a fork light barrier like this: https://cdn-reichelt.de/bilder/web/xxl_ws/A500/ELI_TR8102.png
OP #7436894
Hello Matthias and Wolf,
thank you both very much. This is a very big and great help and motivation for me!
Unfortunately my search was only partially sucessful. What I think I have found is the phototransistor but not the LED, KL 4 and the 68 kOhm resistor.
Here are some photos of presumably Ph 1 and other:
  1. presumably phototransistor Ph 1, img 1
  2. presumably phototransistor Ph 1, img 2
  3. fixation screw of presumably Ph 1
  4. chopper motor and transformator Tr 1, img 1
  5. chopper motor and transformator Tr 1, img 2
  6. transformator 2? pins: I think this cannot be KL 4!
https://workupload.com/archive/Zyh2FN8wUh
(Company: matzetronics) #7436928
Well, that was easy. I marked the forked interrupter in the photo. Its fork contains the LED on one side and the phototransistor in the other. With each turn of the wheel the beam is interrupted for half a turn. On the other half turn it passes and the phototransistor is lit.
From other components of this type i think the output level of the photo TR in your system (approx. 0.2V) is too low. I'd expect more than 2V.
Here's a picture of a CNY36, resembling very closely what you have there: https://www.muekra.de/cny36.html
OP #7437613
Hello Matthias, hello Wolf,
thank you very much. Your help is just great class. You are a good team!
I will order the replacement part and install it.
Secretly I have already suspected that but I did not want to disgrace myself.
Following the thick black cable comming out of the forked interuptor will bring you to KL4 and the 68 Ohm resistor.
As a reminder: *) In newer machines M 1 is directly connected to Tr 1 (not by KL 4)!
The black cable disappears into the "underground" through a hole of the optical system housing but in the next few days (before I get the replacement part) I will try to find KL 4 and the 68-Ohm-resistor, i.e. I will follow the black cable.
https://workupload.com/file/XSMMCaCznDk
After that, I will get back to you and report further.
OP #7451494
Hello,
I am back again. Sorry I was busy with obligatory tasks!
The optocoupler arrived long ago!
I don’t know how the optocoupler is fixed to its mounting and therefore how to remove it. There are two screws, one with a counter nut in direction of the cables, function unknown. Perhaps it is a spacer? When I loosened this screw I could not remove the optocoupler (plastic) or its fixture. The second screw is orthogonal to the first one. I think it only fixes the optocoupler mounting on the metal housing of the chopper-mirror wheel.
https://workupload.com/archive/zRx3HNEa3q
Perhaps the optocoupler may be inserted into the optocoupler mounting so that you have to pull it out backwards to remove it?
Hopefully it is not glued!
What function does the red colour have? Is it used as a marker?
OP #7452687
Hello Wolf and Matthias,
thank you again for reply!
Wolf17 wrote:
Did you check chopper signal lines? Which lines do you mean?
Wolf17 wrote:
Are you sure the optocoupler is defective?
According to Matthias very probably.
Wolf17 wrote:
What is voltage across R12?
I haven’t measured the voltage at 68 kOhm resistor R12 yet. Then I will know if the voltage is U(DC) ≈ 3 V. Before I haven’t measured this voltage I will not change the optocoupler. First off all I have to screw down the photometer housing und then set up the machine because the cable is at the bottom.
Wolf17 wrote:
How much turns per second makes the shutter wheel?
The shutter wheel rotates with net frequency, i.e. f = 50 Hz.
Wolf17 wrote:
Did you measure Ph1 output signal like suggested? Result? Sorry, I can not remember the pin 6 video and it is not available anymore.
Here are the videos of the clock-circuit (CC)-signal measured at pin 6 once again: settings: • 0.2 V/div and 5 ms/div • 0.2 V/div and 2 ms/div • 0.2 V/div and 1 ms/div • 0.5 V/div and 5 ms/div
https://workupload.com/archive/xYtrC25rDq
#7452739
There is a strange ripple each 10ms, please check all supply voltages with oszilloscope: +5V at 68 Ohm R12 and +5V +15V -15V at CC.
I want to know the LED current, what is the voltage across the 68 Ohm R12?
The signal has a roof slope, did you measure with AC coupling instead with DC coupling? I can not see if signal is 0 to 0,25V or for example 0,3 to 0,55V. When making new oszi-video (pictures are enough), please use DC coupling and adjust beam position for zero Volt signal is exact at middle line.
The signal has steep edges, in principle the optocoupler seems to switch, only the 0,25V amplitude is a bit low. Is it 2,5V because you used a 1:10 probe head?
Did you check chopper signal lines? Which lines do you mean? 2023-06-17 15:46 If the clock-circuit is working, there must be the chopper signal on Line 8, 10, 11 and 12 with digital level.
OP #7461929
Hello Wolf,
thank you very much and sorry for my late reply!
Now I have found the 68 Ohm-R 12-resistor (in real a 65 Ohm-resistor!) and KL 4 but with another pin assignment, see photo. Until now I haven't measured the voltage of R 12 and the phototransistor Ph 1 firstly because I don't know if the power unit of the interrupter wheel/chopper (driving shaft or chopper motor) of my photometer will be damaged when measuring while it is standing lateral upright and secondly because I have to know how to measure in detail.
Here is the pin assignment: Left side: resistor R 12 and LED connections? pin 1/socket 1: mass (grey cables are mass (ground?) cables in this machine) pin 2/socket 2: a thin black cable pin 3/socket 3: resistor R 12 (65 Ohm) pin 4/socket 4: resistor R 12 (65 Ohm) and red cable pin 5/socket 5: – pin 6/socket 6: – pin 7/socket 7: – pin 8/socket 8: –
Right side: phototransistor Ph 1 connections pin 1/socket 1: black cable pin 2/socket 2: copper-coloured cable, black curled pin 3/socket 3: copper-coloured cable, red curled pin 4/socket 4: – pin 5/socket 5: – pin 6/socket 6: – pin 7/socket 7: – pin 8/socket 8: –
I suppose that red is plus and black is minus as usually.
Questions:
  1. What is the black cable of pin 1/socket 1 (right side) for?
  2. Is the red cable of pin 4/socket 4 (left side) the supply voltage U = +5 V?
  3. Probe at pin 4/socket 4 (left side) gives the supply voltage of U = +5 V, i.e. before passing resistor R 12 and probe at pin 3/socket 3 should give U ≈ 3 V DC after passing resistor?
  4. How to measure Ph 1? Is it correct, probe at red cable?
Now I know that unfortunately the chopper motor M is not connected by KL 4 (pins 7 and 8) but directly to transformer Tr 1. Therefore, I would have to unsolder a pin of the chopper motor to stop the interrupter wheel without causing a possible damage of the power unit (driving shaft or/and chopper motor) instead of in case of stopping the interrupter wheel by hand!
I have sent following things:
  1. Photos of KL 4 with R 12, pins of LED? and Ph 1 https://workupload.com/archive/5wYc6aYHjT
Setting of probe: always 1:1-position.
I measured with DC coupling, see below!
Both DC switchers were and are in DC position and this time zero line is in middle position!
No, I have always measured in 1:1-position. Once I have measured the signal with that 1:10-setting (without making a photo) but it gave a worse value. After reopening the photometer housing I will make a second measurement together with a photo which I will send then!
clock-circuit: pin 2, supply voltage: U = +5 V pin 16, “ : U = –14.95 V pin 18, “ : U = +14.95 V https://workupload.com/archive/g7Hd2H7PrE
pin 8, clock/chopper/interrupter wheel-signal: U ≈ +3.8 V pin 10, “ : U ≈ +3.8 V pin 11, “ : U ≈ +3.8 V pin 12, “ : U ≈ +3.8 V https://workupload.com/archive/V4SQdcvfdn
pin 4, clock/chopper/interrupter wheel-signal: U ≈ +3.8 V pin 5, “ : U ≈ +3.8 V pin 6, “ : already known, U ≈ 0.4 – 0.6 V pin 7, “ : U ≈ +3.9 V https://workupload.com/archive/J4apHAkbSV
constant-light-protection: pin 24, signal from MSA 1: U ≈ 2 V?, Parkinson-like pin 29: U ≈ 8.8 V instead of U = 8.3 V https://workupload.com/archive/cYHuvhdFLH
energy-control: pin 22, outgoing guide-voltage: U ≈ –4.2 V pin 30, incoming guide-voltage: U ≈ –4.2 V pin 26, supply voltage: U = +5 V pin 27, energy-control voltage: U ≈ 32 V instead of U = 24 V? https://workupload.com/archive/ktt4mm7Xt2
  1. Circuit diagramm of CC https://workupload.com/file/K74apZC5Unk
  2. Photos of CC-circuit board https://workupload.com/archive/JMqQfAQndr
  3. Videos of digital voltmeter counter unit-circuit board (DVC) voltage measurements Photos: https://workupload.com/archive/SCEQFEhMHW Videos: https://workupload.com/archive/VNbbykpejM
I suppose the steady jumping signal voltage of pin 26 and pin 32 should be the same for the other pins. This shows the faulty signal!
  1. Overview circuit diagramm of measurement-value-processing. Follows later! It is needed to (better) understand the videos of 5.
#7462105
Thank you for CC circuit. Even if pin 6 signal look not perfect, the clock-circuit output 4 5 7 and 8 10 11 12 are good. In my opinion, optocoupler and CC digital output signals are working as intended.
OP #7468717
Hello Wolf,
thank you very much!
despite a slight risk of damaging my machine I decided to measure the voltage of La 1 (LED) and signal of Ph 1 (phototransistor) at KL 4.
  1. Supply voltage R 12: U = +5 V
  2. La 1/LED: U ≈ +1.25 V
  3. Signal of Ph 1: U ≈ +0.2 - +0.6 V, rectangular
  4. Signal of CC-6 for comparison
  5. Voltage of R 12: not measured
  6. Current-voltage characteristic of R 12 before LED destruction https://workupload.com/archive/hXud6eQxAE
Because of inattention in measuring the current-voltage characteristic of R 12 I destroyed the LED of the clock. I accidentally forgot to turn off the photometer before measuring resistor R 12! Values after destroying:
  1. Supply voltage of R 12/LED: U = +5 V, unchanged
  2. La 1: U ≈ +4.25 V
  3. Signal of Ph 1: U ≈ +0.6 V, not rectangular any more, i.e. flat line
  4. Voltage of R 12: not measured
  5. Current-voltage characteristic of R 12 after LED destruction https://workupload.com/file/yAah9XGwcfp
In the meantime, I have ordered an original replacement clock to avoid problems and to make sure that everything is working correctly because it is a special design with integrated cables and a special optocoupler (opt.) plastic housing which only fits to its fixture. The two bought cny36-opt. exemplars do not fit anyway! https://workupload.com/archive/Xhu6n9xFn5
As expected all voltages which need a correctly working clock are affected. For example:
  1. CC-8, 10, 11, 12: correct voltage values but now flat lines
  2. CC-29: U = 0 V instead of +8.8 V
  3. CC-10, 22, 30, guide voltage: U = –6.2 V instead of –4 V
  4. DCC-10, 11, guide voltage: U = –6.2 V instead of –4 V
  5. DCC-30, high voltage: U = 0 V instead of –1000 V
Then I took my torch and illuminated the LED, whereupon nearly all values normalized, i.e. flat voltages are digital again, high voltage is U = –700 V and guide voltage is wandering from U = –6.2 V to a less negative value. • CC-6, signal of clock illuminated with torch, 1 V and 1 ms: https://workupload.com/archive/zpQC29G2dy
After installation of the clock and measuring I will further report.
OP #7477425
Hello,
the first repair trial failed. Paradoxically, although everything seems to work correctly it does not work correctly. I don’t know what the problem is at all!
Here are some photos and and a video of the clock LED and clock phototransistor:
  1. Characteristic curve, original defective LED La 1, clock built-out
  2. Characteristic curve, replacement LED, clock built-out
  3. Characteristic curve, original phototransistor Ph 1, clock built-out: the same as 4.
  4. Characteristic curve of replacement phototransistor, clock built-out
  5. Characteristic curve of replacement LED, at KL 4-4/KL 4-2, clock built-in
  6. Supply voltage of resistor R 12/replacement LED at KL 4-5
  7. Forward voltage of replacement LED at KL 4-4
  8. Signal of replacement photoresistor at KL 4-1
  9. Characteristic curve of resistor R 12 at KL 4-5/KL 4-4, clock built-in
https://workupload.com/archive/LgvRSFpGmv
Maybe the resistor is damaged from my faulty measurement. To solve the problem, my idea would be to remove the supply voltage (including resistor R 12) and the optocoupler to detect the fault, isolate it and test CC at the same time. It should be possible to feed in a suitable square wave signal with U = +0.6 V and the correct current using a signal generator. I think that would be the most direct, fastest and best way to proceed.
How does an overdriven phototransistor actually react?